This topic contains a solution. Click here to go to the answer

Author Question: The low-pressure and low-tidal volume alarms are sounding on a mechanically ventilated patient. ... (Read 88 times)

savannahhooper

  • Hero Member
  • *****
  • Posts: 576
The low-pressure and low-tidal volume alarms are sounding on a mechanically ventilated patient. Measurement of the cuff pressure reveals 18 cm H2O. What action should be taken?
 
  a. Replace the endotracheal tube with a larg-er size.
  b. Add enough air to the cuff to maintain the cuff pressure at 34 cm H2O.
  c. Increase the set pressure to increase the tidal volume and compensate for the leak.
  d. Add air until a slight leak is heard while auscultating the larynx, then measure pressure.

Question 2

A patient's transairway pressure (PTA) is rising while the plateau pressure (Pplateau) remains un-changed. The treatment plan that could correct this problem includes which of the following?
 
  1. Administer a bronchodilator.
  2. Insert a chest tube.
  3. Measure unintended positive end-expiratory pressure (auto-PEEP).
  4. Suction airway secretions.
  a. 2 only
  b. 2 and 4 only
  c. 1 and 4 only
  d. 1 and 3 only



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

deja

  • Sr. Member
  • ****
  • Posts: 332
Answer to Question 1

ANS: D
The minimum leak technique (MLT) should be used whenever possible to avoid tracheal necrosis associated with cuff overinflation. The size of the tube and patient is not given, so one cannot say whether or not the endotracheal (ET) tube is too small. 34 cm H2O is the maximum pressure that should be in a cuff. However, since all airways are different, the most effective way to minimize the risk of tracheal problems is to use MLT.

Answer to Question 2

ANS: C
Airway resistance can be estimated for a ventilated patient using transairway pressure (PTA). When the PTA rises, the airway resistance is rising. When plateau pressure (Pplateau) is constant, this means that the static compliance has not changed. In this question the PTA is rising while the Pplat-eau remains the same. This is due to an increase in Raw and can be corrected by administering a bronchodilator and/or suctioning the airways. An increase in Pplateau could occur from a pneumo-thorax, which would require the insertion of a chest tube and also from dynamic hyperinflation, which can be measured as unintended positive end-expiratory pressure (auto-PEEP).




savannahhooper

  • Member
  • Posts: 576
Reply 2 on: Jul 16, 2018
:D TYSM


  • Member
  • Posts:
Reply 3 on: Yesterday
Gracias!

 

Did you know?

Methicillin-resistant Staphylococcus aureus or MRSA was discovered in 1961 in the United Kingdom. It if often referred to as a superbug. MRSA infections cause more deaths in the United States every year than AIDS.

Methicilli ...
Did you know?

About 100 new prescription or over-the-counter drugs come into the U.S. market every year.

Did you know?

Autoimmune diseases occur when the immune system destroys its own healthy tissues. When this occurs, white blood cells cannot distinguish between pathogens and normal cells.

Did you know?

As the western states of America were settled, pioneers often had to drink rancid water from ponds and other sources. This often resulted in chronic diarrhea, causing many cases of dehydration and death that could have been avoided if clean water had been available.

Did you know?

Barbituric acid, the base material of barbiturates, was first synthesized in 1863 by Adolph von Bayer. His company later went on to synthesize aspirin for the first time, and Bayer aspirin is still a popular brand today.

For a complete list of videos, visit our video library