This topic contains a solution. Click here to go to the answer

Author Question: During high-frequency oscillation ventilation (HFOV), how is fresh gas supplied with the ventilator ... (Read 87 times)

Tirant22

  • Hero Member
  • *****
  • Posts: 532
During high-frequency oscillation ventilation (HFOV), how is fresh gas supplied with the ventilator circuit?
 
  A. through an oscillator
  B. through an entrainment port influenced by the oscillation rate
  C. via bias flow
  D. by the size of the ETT

Question 2

The therapist is about to ventilate a neonate with HFJV. She has different injectors and can place the one she selects in a variety of locations. Where should the therapist position the injector in relation to the patient's airway to minimize the patient's functional dead space?
 
  A. at the external opening of the ETT
  B. at the distal tip of the ETT
  C. in the breathing circuit of the ventilator
  D. beyond the distal tip of the ETT



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

orangecrush

  • Sr. Member
  • ****
  • Posts: 350
Answer to Question 1

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: During HFOV, fresh gas is supplied with the ventilator circuit as a bias flow. The mean airway pressure is adjusted by the relationship between fresh gas inflow and any positive or negative pressure placed on the gas outflow from the bias flow circuit.
D. Incorrect response: See explanation C.

Answer to Question 2

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: High-frequency jets operate on the principle of a nozzle or an injector that creates a high velocity jet of gas directed into the patient's lungs. These injectors are usually 1 to 3 cm in diameter, and can be placed in one of several locations in the ventilator circuit and/or patient airway. Generally the further down the airway the injector is placed, the less is the patient's functional dead space.





 

Did you know?

In the United States, congenital cytomegalovirus causes one child to become disabled almost every hour. CMV is the leading preventable viral cause of development disability in newborns. These disabilities include hearing or vision loss, and cerebral palsy.

Did you know?

There are immediate benefits of chiropractic adjustments that are visible via magnetic resonance imaging (MRI). It shows that spinal manipulation therapy is effective in decreasing pain and increasing the gaps between the vertebrae, reducing pressure that leads to pain.

Did you know?

More than 34,000 trademarked medication names and more than 10,000 generic medication names are in use in the United States.

Did you know?

The Centers for Disease Control and Prevention has released reports detailing the deaths of infants (younger than 1 year of age) who died after being given cold and cough medications. This underscores the importance of educating parents that children younger than 2 years of age should never be given over-the-counter cold and cough medications without consulting their physicians.

Did you know?

No drugs are available to relieve parathyroid disease. Parathyroid disease is caused by a parathyroid tumor, and it needs to be removed by surgery.

For a complete list of videos, visit our video library