Author Question: The therapist is delivering assist-control ventilation providing a constant VT to a patient ... (Read 82 times)

armygirl

  • Hero Member
  • *****
  • Posts: 556
The therapist is delivering assist-control ventilation providing a constant VT to a patient experiencing ineffective triggering. The therapist also notes that expiratory flow fails to return to baseline before the initiation of the next mandatory breath. What should the therapist do at this time to correct this problem?
 
  A. Increase the inspiratory flow.
  B. Institute an inspiratory pause.
  C. Conduct an expiratory pause maneuver.
  D. Shorten the expiratory time.

Question 2

As the therapist works with a mechanically ventilated patient, he notices that the ventilator is auto-triggering. What should the therapist do to eliminate this problem? I. Add water to the heated humidifier. II. Check the patient-ventilator system for leaks. III. Scrutinize the breathing circuit for the accumulation of condensate. IV. Ensure that the high-pressure alarm is appropriately set
 
  A. I, II only
  B. II, III only
  C. I, III, IV only
  D, I, II, IV only



cpetit11

  • Sr. Member
  • ****
  • Posts: 321
Answer to Question 1

ANS: A
A. Correct response: This patient is receiving assist-control ventilation with a constant tidal volume and is experiencing ineffective triggering. By increasing the inspiratory flow, the expiratory time lengthens, and the dynamic hyperinflation and auto-PEEP decrease. Otherwise, the patient would be required to develop subatmospheric pressure equivalent to the sum of the applied PEEP and auto-PEEP to trigger inspiration.
B. Incorrect response: Instituting an inspiratory pause would aggravate the problem by lengthening inspiratory time further and shortening expiratory time. This situation would worsen the dynamic hyperinflation.
C. Incorrect response: Performing an expiratory pause maneuver would not rectify this situation.
D. Incorrect response: Shortening the expiratory time would worsen the problem by causing further dynamic hyperinflation.

Answer to Question 2

ANS: B
A. Incorrect response: See explanation B.
B. Correct response: Auto-triggering occurs when the triggering system of the ventilator is overly sensitive. Factors that can cause auto-triggering include condensate in the ventilator breathing circuit and a leak in the patient-ventilator system.
Respiratory system depression associated with a low respiratory rate is susceptible to auto-triggering. This condition prevents flow for a certain period. Zero flow through the system for an extended time makes the system vulnerable to triggering from changes in the airway pressure occurring within the ventilator breathing circuit. These vulnerabilities are not related to patient effort;, rather they pertain to pressure changes within the circuit, even an event as subtle as cardiogenic oscillations.
C. Incorrect response: See explanation B.
D. Incorrect response: See explanation B.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

Most women experience menopause in their 50s. However, in 1994, an Italian woman gave birth to a baby boy when she was 61 years old.

Did you know?

The Centers for Disease Control and Prevention has released reports detailing the deaths of infants (younger than 1 year of age) who died after being given cold and cough medications. This underscores the importance of educating parents that children younger than 2 years of age should never be given over-the-counter cold and cough medications without consulting their physicians.

Did you know?

The use of salicylates dates back 2,500 years to Hippocrates's recommendation of willow bark (from which a salicylate is derived) as an aid to the pains of childbirth. However, overdosage of salicylates can harm body fluids, electrolytes, the CNS, the GI tract, the ears, the lungs, the blood, the liver, and the kidneys and cause coma or death.

Did you know?

Asthma attacks and symptoms usually get started by specific triggers (such as viruses, allergies, gases, and air particles). You should talk to your doctor about these triggers and find ways to avoid or get rid of them.

Did you know?

Aspirin is the most widely used drug in the world. It has even been recognized as such by the Guinness Book of World Records.

For a complete list of videos, visit our video library