Author Question: An orally intubated patient is receiving PC-CMV. The therapist suddenly notices that the patient's ... (Read 42 times)

elizabeth18

  • Hero Member
  • *****
  • Posts: 559
An orally intubated patient is receiving PC-CMV. The therapist suddenly notices that the patient's exhaled tidal volume has fallen significantly. Which of the following occurrences is the likely cause of this loss of volume?
 
  A. The cuff on the endotracheal tube is overinflated.
  B. The patient's static compliance has increased.
  C. An oversized endotracheal tube has been used.
  D. The endotracheal tube has migrated into the right mainstem bronchus.

Question 2

The diagram show below depicts an end-expiratory valve used to generate PEEP during mechanical ventilation. Which of the following statements describes the function of this valve?
 
  A. A decreased flow through this apparatus causes an increased PEEP.
  B. An increased flow through this device produces an increased PEEP.
  C. When the flow through this apparatus remains constant, the PEEP level generated fluctuates.
  D. The PEEP level produced by this device is proportional to the level of flow.



sarah_brady415

  • Sr. Member
  • ****
  • Posts: 328
Answer to Question 1

ANS: D
A. Incorrect response: An overinflated endotracheal tube cuff will not affect the patient's tidal volume.
B. Incorrect response: If static compliance increases, tidal volume will increase in PC-CMV. Tidal volume will decrease when Raw increases in PC-CMV.
C. Incorrect response: Using an oversized endotracheal tube will not cause the patient's tidal volume to drop suddenly.
D. Correct response: Commonly, distal migration of the endotracheal tube is detected by an increase in airway resistance during volume control ventilation, and a decrease in the tidal volume during pressure control ventilation.

Answer to Question 2

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: The device pictured in response D is an end-expiratory valve called a flow resistor used for generating PEEP. A flow resistor requires a constant flow to maintain a constant pressure. Therefore, if the flow through a flow resistor is low, the PEEP level developed will be low. Conversely, when the flow through this device is high, the level of PEEP will be high. Another type of end-expiratory valve is the threshold resistor, which maintains a steady level of PEEP regardless of the flow moving through the device.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

Drugs are in development that may cure asthma and hay fever once and for all. They target leukotrienes, which are known to cause tightening of the air passages in the lungs and increase mucus productions in nasal passages.

Did you know?

Computer programs are available that crosscheck a new drug's possible trade name with all other trade names currently available. These programs detect dangerous similarities between names and alert the manufacturer of the drug.

Did you know?

Although the Roman numeral for the number 4 has always been taught to have been "IV," according to historians, the ancient Romans probably used "IIII" most of the time. This is partially backed up by the fact that early grandfather clocks displayed IIII for the number 4 instead of IV. Early clockmakers apparently thought that the IIII balanced out the VIII (used for the number 8) on the clock face and that it just looked better.

Did you know?

Oxytocin is recommended only for pregnancies that have a medical reason for inducing labor (such as eclampsia) and is not recommended for elective procedures or for making the birthing process more convenient.

Did you know?

To combat osteoporosis, changes in lifestyle and diet are recommended. At-risk patients should include 1,200 to 1,500 mg of calcium daily either via dietary means or with supplements.

For a complete list of videos, visit our video library