An orally intubated patient is receiving PC-CMV. The therapist suddenly notices that the patient's exhaled tidal volume has fallen significantly. Which of the following occurrences is the likely cause of this loss of volume?
 
  A. The cuff on the endotracheal tube is overinflated.
  B. The patient's static compliance has increased.
  C. An oversized endotracheal tube has been used.
  D. The endotracheal tube has migrated into the right mainstem bronchus.

Question 2

The diagram show below depicts an end-expiratory valve used to generate PEEP during mechanical ventilation. Which of the following statements describes the function of this valve?
 
  A. A decreased flow through this apparatus causes an increased PEEP.
  B. An increased flow through this device produces an increased PEEP.
  C. When the flow through this apparatus remains constant, the PEEP level generated fluctuates.
  D. The PEEP level produced by this device is proportional to the level of flow.



sarah_brady415

  • Sr. Member
  • ****
  • Posts: 328
Answer to Question 1

ANS: D
A. Incorrect response: An overinflated endotracheal tube cuff will not affect the patient's tidal volume.
B. Incorrect response: If static compliance increases, tidal volume will increase in PC-CMV. Tidal volume will decrease when Raw increases in PC-CMV.
C. Incorrect response: Using an oversized endotracheal tube will not cause the patient's tidal volume to drop suddenly.
D. Correct response: Commonly, distal migration of the endotracheal tube is detected by an increase in airway resistance during volume control ventilation, and a decrease in the tidal volume during pressure control ventilation.

Answer to Question 2

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: The device pictured in response D is an end-expiratory valve called a flow resistor used for generating PEEP. A flow resistor requires a constant flow to maintain a constant pressure. Therefore, if the flow through a flow resistor is low, the PEEP level developed will be low. Conversely, when the flow through this device is high, the level of PEEP will be high. Another type of end-expiratory valve is the threshold resistor, which maintains a steady level of PEEP regardless of the flow moving through the device.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

Fungal nail infections account for up to 30% of all skin infections. They affect 5% of the general population—mostly people over the age of 70.

Did you know?

When intravenous medications are involved in adverse drug events, their harmful effects may occur more rapidly, and be more severe than errors with oral medications. This is due to the direct administration into the bloodstream.

Did you know?

There are 20 feet of blood vessels in each square inch of human skin.

Did you know?

The Romans did not use numerals to indicate fractions but instead used words to indicate parts of a whole.

Did you know?

Medication errors are more common among seriously ill patients than with those with minor conditions.

For a complete list of videos, visit our video library