Author Question: Mean airway pressure can be increased when ventilating with the Bunnell Life Pulse High-Frequency ... (Read 77 times)

storky111

  • Hero Member
  • *****
  • Posts: 561
Mean airway pressure can be increased when ventilating with the Bunnell Life Pulse High-Frequency Jet Ventilator by doing which of the following?
 
  a. Closing off the purge valve
  b. Decreasing the rate on the Bunnell
  c. Decreasing the positive end-expiratory pressure on the Bunnell
  d. Increasing rate and volume of sigh breaths on the conventional ventilator

Question 2

Which of the following statements concerning the volume guarantee mode on the Drger Baby-log 8000 ventilator is true?
 
  a. The breaths are volume-limited in this mode.
  b. This mode does not operate with as-sist/control.
  c. The Drger Babylog 8000 and 8000+ both have this mode.
  d. Positional endotracheal tube leaks may be hazardous with this mode.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

kbennett34

  • Sr. Member
  • ****
  • Posts: 333
Answer to Question 1

ANS: D
The function of the purge valve is to provide a moisture-free environment in the monitoring line of the hi-lo jet tube. Decreasing the rate decreases the mean airway pressure because less time is spent at the inspiratory pressure. Decreasing the positive end-expiratory pressure level decreases the mean airway pressure. Increasing the rate and volume of the sigh breaths increases the mean airway pressure.

Answer to Question 2

ANS: D
When the volume guarantee mode is activated, the ventilator continuously measures and com-pares inspired and expired tidal volume. Each following breath uses the comparisons from the previous breath to make adjustments to the peak inspiratory pressure so that a volume as close as possible to the preset value can be delivered. If there is a positional leak and it is active, the ex-pired tidal volume will be lower than the inspired tidal volume, and the ventilator will increase the pressure to attempt to match the two monitored volumes. This increase in pressure continues if the leak continues. However, if the patient's position changes and the leak is not occurring, the breath delivered will be significantly larger than intended. The limit to the amount of volume is 130 of the targeted volume. At this point, the exhalation valve will open, and no additional ventilator-driven gas will be delivered to the patient. Close monitoring of patient-ventilator in-teraction is needed during use of this mode.




storky111

  • Member
  • Posts: 561
Reply 2 on: Jul 16, 2018
Thanks for the timely response, appreciate it


Jsherida

  • Member
  • Posts: 314
Reply 3 on: Yesterday
:D TYSM

 

Did you know?

Certain chemicals, after ingestion, can be converted by the body into cyanide. Most of these chemicals have been removed from the market, but some old nail polish remover, solvents, and plastics manufacturing solutions can contain these substances.

Did you know?

Barbituric acid, the base material of barbiturates, was first synthesized in 1863 by Adolph von Bayer. His company later went on to synthesize aspirin for the first time, and Bayer aspirin is still a popular brand today.

Did you know?

There are 60,000 miles of blood vessels in every adult human.

Did you know?

Computer programs are available that crosscheck a new drug's possible trade name with all other trade names currently available. These programs detect dangerous similarities between names and alert the manufacturer of the drug.

Did you know?

In 2006, a generic antinausea drug named ondansetron was approved. It is used to stop nausea and vomiting associated with surgery, chemotherapy, and radiation therapy.

For a complete list of videos, visit our video library