Author Question: The tubing that brings the lyse reagent to the hemoglobin cuvette on an automated cell counter is ... (Read 138 times)

abarnes

  • Hero Member
  • *****
  • Posts: 556
The tubing that brings the lyse reagent to the hemoglobin cuvette on an automated cell counter is pinched and not delivering any reagent. All hemoglobin values are greater than 20 g/dL. This represents what type of error?
 
  a. Random
  b. Imprecision
  c. Constant systematic
  d. Proportional systematic

Question 2

One of two controls that have been evaluated over the last 28 days gives a result on day 29 between 2 and 3 SDs of the mean; the other control is within 2 SDs of its mean. What is the correct procedure to follow?
 
  a. Ignore the result unless it happens again the next day.
  b. Rerun the control and if acceptable continue with patients.
  c. Recalibrate the instrument.
  d. Open new vials of controls and repeat both controls.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

AmberC1996

  • Sr. Member
  • ****
  • Posts: 317
Answer to Question 1

ANS: C
A constant systematic error is one in which the magnitude of the error remains the same throughout the range of the test measurement. The error is proportional if the magnitude varies relative to the result. This is not a random error, which happens only infrequently and is not predictable. Precision requires multiple measurements of the same specimen and evaluates the ability to consistently reproduce the result.

Answer to Question 2

ANS: B
One control is acceptable, whereas the other is a warning that the method may be going out of control. The test option in this case is to repeat the analysis of the control, and if it is acceptable, then continue with patient analysis, reporting the results. The instrument does not appear to need recalibration because one control is acceptable and the other is within 3 SDs (1 result of 20 can acceptably be within 3 SDs). If the repeat on the out of control vial is still out between 2 and 3 SDs, then a new vial of that control should be opened and analyzed. The control that was acceptable does need to be repeated.




abarnes

  • Member
  • Posts: 556
Reply 2 on: Jun 25, 2018
Thanks for the timely response, appreciate it


sarah_brady415

  • Member
  • Posts: 328
Reply 3 on: Yesterday
YES! Correct, THANKS for helping me on my review

 

Did you know?

Asthma cases in Americans are about 75% higher today than they were in 1980.

Did you know?

The B-complex vitamins and vitamin C are not stored in the body and must be replaced each day.

Did you know?

Thyroid conditions may make getting pregnant impossible.

Did you know?

Human neurons are so small that they require a microscope in order to be seen. However, some neurons can be up to 3 feet long, such as those that extend from the spinal cord to the toes.

Did you know?

The lipid bilayer is made of phospholipids. They are arranged in a double layer because one of their ends is attracted to water while the other is repelled by water.

For a complete list of videos, visit our video library