  • Posts: 556
The tubing that brings the lyse reagent to the hemoglobin cuvette on an automated cell counter is pinched and not delivering any reagent. All hemoglobin values are greater than 20 g/dL. This represents what type of error?
 
  a. Random
  b. Imprecision
  c. Constant systematic
  d. Proportional systematic

Question 2

One of two controls that have been evaluated over the last 28 days gives a result on day 29 between 2 and 3 SDs of the mean; the other control is within 2 SDs of its mean. What is the correct procedure to follow?
 
  a. Ignore the result unless it happens again the next day.
  b. Rerun the control and if acceptable continue with patients.
  c. Recalibrate the instrument.
  d. Open new vials of controls and repeat both controls.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

AmberC1996

  • Sr. Member
  • ****
  • Posts: 317
Answer to Question 1

ANS: C
A constant systematic error is one in which the magnitude of the error remains the same throughout the range of the test measurement. The error is proportional if the magnitude varies relative to the result. This is not a random error, which happens only infrequently and is not predictable. Precision requires multiple measurements of the same specimen and evaluates the ability to consistently reproduce the result.

Answer to Question 2

ANS: B
One control is acceptable, whereas the other is a warning that the method may be going out of control. The test option in this case is to repeat the analysis of the control, and if it is acceptable, then continue with patient analysis, reporting the results. The instrument does not appear to need recalibration because one control is acceptable and the other is within 3 SDs (1 result of 20 can acceptably be within 3 SDs). If the repeat on the out of control vial is still out between 2 and 3 SDs, then a new vial of that control should be opened and analyzed. The control that was acceptable does need to be repeated.




abarnes

  • Member
  • Posts: 556
Reply 2 on: Jun 25, 2018
Great answer, keep it coming :)


Sarahjh

  • Member
  • Posts: 370
Reply 3 on: Yesterday
Wow, this really help

 

Did you know?

During the twentieth century, a variant of the metric system was used in Russia and France in which the base unit of mass was the tonne. Instead of kilograms, this system used millitonnes (mt).

Did you know?

To combat osteoporosis, changes in lifestyle and diet are recommended. At-risk patients should include 1,200 to 1,500 mg of calcium daily either via dietary means or with supplements.

Did you know?

Individuals are never “cured” of addictions. Instead, they learn how to manage their disease to lead healthy, balanced lives.

Did you know?

In 2012, nearly 24 milliion Americans, aged 12 and older, had abused an illicit drug, according to the National Institute on Drug Abuse (NIDA).

Did you know?

Vital signs (blood pressure, temperature, pulse rate, respiration rate) should be taken before any drug administration. Patients should be informed not to use tobacco or caffeine at least 30 minutes before their appointment.

For a complete list of videos, visit our video library