Answer to Question 1
ANS: B
A. Incorrect response: Instituting airway pressure release ventilation at this time seems incongruous because these changing inspiratory demands caused by periods of tachypnea appear transient. No apparent need exists to establish two levels of supra-atmospheric pressure now.
B. Correct response: Automatic tube compensation (ATC) uses the known resistive characteristics of the patient's artificial airway to overcome the imposed work of breathing (WOB) caused by that airway. ATC is not a mode of ventilation. It is merely an add-on designed to eliminate the resistive characteristics of the artificial airway and the flow demand experienced by the patient. Choosing one level of PS is inadequate to meet the changing demands caused by variable inspiratory flows during tachypneic periods.
C. Incorrect response: MMV enables the patient to breathe spontaneously while, at the same time, ensuring that the patient receives a minimum minute ventilation.
D. Incorrect response: Inverse ratio ventilation would be a drastic measure at this time. No indication exists for establishing a prolonged TI with a shortened TE, which would elevate this patient's mean airway pressure.
Answer to Question 2
ANS: B
A. Incorrect response: Volume support (VS) ventilation is pressure support ventilation employing volume as the feedback control for continuously adjusting the pressure support level. Like pressure support ventilation, VS is entirely a spontaneous breathing mode of ventilation.
B. Correct response: A volume support breath is a volume targeted breath that adjusts the pressure over several breaths to achieve a set VT. Therefore, if the volume exiting the ventilator does not equal the set VT, the pressure will increase to raise the VT. On the other hand, if the volume delivered is too large, the pressure will decrease.
If the target tidal volume (minute ventilation) is not met, these volume adjustments are accomplished by the ventilator over the course of three successive breaths. The volume added will not permit the pressure to increase by more than 3 cm H2O for each consecutive breath. The targeted VT will increase up to 150 of the set VT, thus maintaining a near constant minute ventilation.
C. Incorrect response: No backup respiratory rate exists for the VS mode. In fact, the respiratory rate is preset on the ventilator. So, when the minute ventilation decreases, the VT is adjusted to maintain a fairly constant minute ventilation.
D. Incorrect response: Inspiratory time will not be adjusted to achieve the targeted VT.