Author Question: While the therapist provides HFJV to a neonate, which of the following factors serve as the basis ... (Read 36 times)

james9437

  • Hero Member
  • *****
  • Posts: 568
While the therapist provides HFJV to a neonate, which of the following factors serve as the basis for establishing ventilator settings? I. peak and airway pressure monitoring II. arterial blood gas data III. capnography data IV. visual inspection of the neonate's thorax
 
  A. I, II only
  B. II, III only
  C. I, II, III only
  D. I, II, IV only

Question 2

The therapist is setting up an INOvent to administer inhaled NO to a neonatal patient. The therapist wants to deliver a concentration of 20 ppm of NO to this patient. What must the therapist do to the INOvent before attaching it to the patient?
 
  A. Disconnect the INOvent from the Y-piece in the inspiratory breathing circuit before turning on the flowmeter.
  B. Turn off the following alarms before setting the O2 flow and the injector: loss of gas pressure, calibration required, and delivery system failure.
  C. Set an O2 flow of 15 L/minute and squeeze the bag four times to clear residual NO2.
  D. Activate the dual channels to control NO delivery and to monitor the INOvent output.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

meryzewe

  • Sr. Member
  • ****
  • Posts: 346
Answer to Question 1

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: Ventilation settings used with jet ventilators are usually established according to peak and mean airway pressure monitoring, visual inspection of chest movement, and arterial blood gas data.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: A manual NO delivery system is provided by the INOvent delivery system. With an O2 flow to the manual ventilator set at 15 L/minute, INOvent injects gas to provide an NO concentration of 20 ppm. As with any manual ventilator system for NO, the bag should be squeezed three to five times to clear residual NO2 before the ventilator is attached to the patient.
D. Incorrect response: See explanation C.




james9437

  • Member
  • Posts: 568
Reply 2 on: Jul 16, 2018
:D TYSM


dreamfighter72

  • Member
  • Posts: 355
Reply 3 on: Yesterday
Great answer, keep it coming :)

 

Did you know?

For high blood pressure (hypertension), a new class of drug, called a vasopeptidase blocker (inhibitor), has been developed. It decreases blood pressure by simultaneously dilating the peripheral arteries and increasing the body's loss of salt.

Did you know?

About 3% of all pregnant women will give birth to twins, which is an increase in rate of nearly 60% since the early 1980s.

Did you know?

Astigmatism is the most common vision problem. It may accompany nearsightedness or farsightedness. It is usually caused by an irregularly shaped cornea, but sometimes it is the result of an irregularly shaped lens. Either type can be corrected by eyeglasses, contact lenses, or refractive surgery.

Did you know?

Many people have small pouches in their colons that bulge outward through weak spots. Each pouch is called a diverticulum. About 10% of Americans older than age 40 years have diverticulosis, which, when the pouches become infected or inflamed, is called diverticulitis. The main cause of diverticular disease is a low-fiber diet.

Did you know?

Certain chemicals, after ingestion, can be converted by the body into cyanide. Most of these chemicals have been removed from the market, but some old nail polish remover, solvents, and plastics manufacturing solutions can contain these substances.

For a complete list of videos, visit our video library