Author Question: While checking a polarographic (Clark) electrode, you determine that the device fails to read 100 ... (Read 21 times)

vicky

  • Hero Member
  • *****
  • Posts: 586
While checking a polarographic (Clark) electrode, you determine that the device fails to read 100 when exposed to pure O2. Which of the following actions would be the proper first step?
 
  a. Check the silica crystals.
  b. Send the device out for repair.
  c. Replace the analyzer's batteries.
  d. Change the analyzer's fuel cell.

Question 2

During calibration of a Clark polarographic O2 analyzer, you cannot get the sensor to read 100 when exposed to 100 O2, even after adjusting its calibration control. The unit has new batteries.
 
  Which of the following actions would be correct?
  a. Use it only with low FiO2 values.
  b. Send the device out for repair.
  c. Replace the unit's batteries.
  d. Change the sensor or electrode.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

jliusyl

  • Sr. Member
  • ****
  • Posts: 328
Answer to Question 1

ANS: C
Because a low battery condition is so common with Clark electrode systems, the first step in troubleshooting is to replace the batteries.

Answer to Question 2

ANS: D
If the analyzer still does not calibrate on fresh batteries, the problem is probably a depleted sen-sor. With most analyzers, a depleted sensor must be replaced (some Clark electrodes can be re-charged).




vicky

  • Member
  • Posts: 586
Reply 2 on: Jul 16, 2018
Gracias!


mcabuhat

  • Member
  • Posts: 344
Reply 3 on: Yesterday
Thanks for the timely response, appreciate it

 

Did you know?

Methicillin-resistant Staphylococcus aureus or MRSA was discovered in 1961 in the United Kingdom. It if often referred to as a superbug. MRSA infections cause more deaths in the United States every year than AIDS.

Methicilli ...
Did you know?

Before a vaccine is licensed in the USA, the Food and Drug Administration (FDA) reviews it for safety and effectiveness. The CDC then reviews all studies again, as well as the American Academy of Pediatrics and the American Academy of Family Physicians. Every lot of vaccine is tested before administration to the public, and the FDA regularly inspects vaccine manufacturers' facilities.

Did you know?

Drying your hands with a paper towel will reduce the bacterial count on your hands by 45–60%.

Did you know?

The modern decimal position system was the invention of the Hindus (around 800 AD), involving the placing of numerals to indicate their value (units, tens, hundreds, and so on).

Did you know?

Pope Sylvester II tried to introduce Arabic numbers into Europe between the years 999 and 1003, but their use did not catch on for a few more centuries, and Roman numerals continued to be the primary number system.

For a complete list of videos, visit our video library