Author Question: While checking a polarographic (Clark) electrode, you determine that the device fails to read 100 ... (Read 39 times)

vicky

  • Hero Member
  • *****
  • Posts: 586
While checking a polarographic (Clark) electrode, you determine that the device fails to read 100 when exposed to pure O2. Which of the following actions would be the proper first step?
 
  a. Check the silica crystals.
  b. Send the device out for repair.
  c. Replace the analyzer's batteries.
  d. Change the analyzer's fuel cell.

Question 2

During calibration of a Clark polarographic O2 analyzer, you cannot get the sensor to read 100 when exposed to 100 O2, even after adjusting its calibration control. The unit has new batteries.
 
  Which of the following actions would be correct?
  a. Use it only with low FiO2 values.
  b. Send the device out for repair.
  c. Replace the unit's batteries.
  d. Change the sensor or electrode.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

jliusyl

  • Sr. Member
  • ****
  • Posts: 328
Answer to Question 1

ANS: C
Because a low battery condition is so common with Clark electrode systems, the first step in troubleshooting is to replace the batteries.

Answer to Question 2

ANS: D
If the analyzer still does not calibrate on fresh batteries, the problem is probably a depleted sen-sor. With most analyzers, a depleted sensor must be replaced (some Clark electrodes can be re-charged).




vicky

  • Member
  • Posts: 586
Reply 2 on: Jul 16, 2018
Excellent


vickyvicksss

  • Member
  • Posts: 351
Reply 3 on: Yesterday
Thanks for the timely response, appreciate it

 

Did you know?

There are more nerve cells in one human brain than there are stars in the Milky Way.

Did you know?

Every 10 seconds, a person in the United States goes to the emergency room complaining of head pain. About 1.2 million visits are for acute migraine attacks.

Did you know?

Each year in the United States, there are approximately six million pregnancies. This means that at any one time, about 4% of women in the United States are pregnant.

Did you know?

When intravenous medications are involved in adverse drug events, their harmful effects may occur more rapidly, and be more severe than errors with oral medications. This is due to the direct administration into the bloodstream.

Did you know?

The most destructive flu epidemic of all times in recorded history occurred in 1918, with approximately 20 million deaths worldwide.

For a complete list of videos, visit our video library