Author Question: If a pressure-support breath on a CareFusion V.I.P. Bird Infant-Pediatric Ventilator is 25 mL and ... (Read 57 times)

Engineer

  • Hero Member
  • *****
  • Posts: 527
If a pressure-support breath on a CareFusion V.I.P. Bird Infant-Pediatric Ventilator is 25 mL and the peak inspiratory flow is 10 L/min, the breath will flow-cycle at _____ L/min.
 
  a. 0.50 c. 1.25
  b. 1.00 d. 2.00

Question 2

A pressure-support breath on a CareFusion V.I.P. Bird delivers 230 mL, and the peak inspiratory flow is 18 L/min. This breath will flow-cycle at _____ L/min.
 
  a. 0.9 c. 3.6
  b. 1.8 d. 4.5



vseab

  • Sr. Member
  • ****
  • Posts: 323
Answer to Question 1

ANS: A
Between tidal volumes of 0 mL and 50 mL, the flow-cycle value is 5 of peak flow. Therefore, 5  10 L/min = 0.5 L/min.

Answer to Question 2

ANS: D
For delivered volumes greater than 200 mL, the flow cycle value is 25 of peak flow. Therefore, 25  18 L/min = 4.5 L/min.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

The ratio of hydrogen atoms to oxygen in water (H2O) is 2:1.

Did you know?

Today, nearly 8 out of 10 pregnant women living with HIV (about 1.1 million), receive antiretrovirals.

Did you know?

Limit intake of red meat and dairy products made with whole milk. Choose skim milk, low-fat or fat-free dairy products. Limit fried food. Use healthy oils when cooking.

Did you know?

Cocaine was isolated in 1860 and first used as a local anesthetic in 1884. Its first clinical use was by Sigmund Freud to wean a patient from morphine addiction. The fictional character Sherlock Holmes was supposed to be addicted to cocaine by injection.

Did you know?

To prove that stomach ulcers were caused by bacteria and not by stress, a researcher consumed an entire laboratory beaker full of bacterial culture. After this, he did indeed develop stomach ulcers, and won the Nobel Prize for his discovery.

For a complete list of videos, visit our video library