Author Question: The therapist is about to initiate mechanical ventilation using heliox as the source gas. Which of ... (Read 45 times)

jilianpiloj

  • Hero Member
  • *****
  • Posts: 521
The therapist is about to initiate mechanical ventilation using heliox as the source gas. Which of the following considerations should the therapist take into account while setting the mechanical ventilator?
 
  A. to use volume ventilation
  B. to employ pressure ventilation
  C. to administer continuous mandatory ventilation with the patient sedated and paralyzed
  D. to avoid using any closed-loop modes of mechanical ventilation

Question 2

The therapist is using a Servo ventilator to ventilate a patient with heliox. How should the therapist anticipate the flow sensors on this ventilator to be affected by heliox?
 
  A. Because the flow sensors operate on the principle of thermal conductivity, they will be affected by heliox.
  B. Because the flow sensors are influenced by viscosity, they will be less affected by heliox.
  C. They are affected by gas density and will read erroneously.
  D. They will sense the flow of heliox accurately.



zhanghao

  • Sr. Member
  • ****
  • Posts: 316
Answer to Question 1

ANS: B
A. Incorrect response: See explanation B.
B. Correct response: Heliox administration via mechanical ventilation can be problematic. Ventilators are designed to deliver a mixture of air and O2. The different density and viscosity of helium can affect the delivered Vt and the exhaled Vt. For some ventilators, no reliable Vt is delivered with heliox. For other ventilators, a Vt higher than desired may be delivered. This problem can be circumvented partially by selecting pressure ventilation rather than volume ventilation. Unlike flow sensors, pressure sensors are not affected by a different gas composition.
C. Incorrect response: See explanation B.
D. Incorrect response: See explanation B.

Answer to Question 2

ANS: B
A. Incorrect response: See explanation B.
B. Correct response: The effect of heliox on the ability of the ventilator to correctly monitor flow and Vt depends on the method that is used for this measurement. Monitoring devices that are density dependent are inaccurate in the presence of heliox. Devices that use the principle of thermal conductivity are also affected. However, devices that are influenced by gas viscosity rather than by gas density are affected to a lesser degree because the viscosity of helium is only slightly different from that of air or O2. Screen pneumotachometers such as those used in the Servo ventilators are affected by viscosity rather than density. Thus, the flow sensors in these ventilators are affected to a lesser degree than occurs in other ventilators.
C. Incorrect response: See explanation B.
D. Incorrect response: See explanation B.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

Certain rare plants containing cyanide include apricot pits and a type of potato called cassava. Fortunately, only chronic or massive ingestion of any of these plants can lead to serious poisoning.

Did you know?

The effects of organophosphate poisoning are referred to by using the abbreviations “SLUD” or “SLUDGE,” It stands for: salivation, lacrimation, urination, defecation, GI upset, and emesis.

Did you know?

Although not all of the following muscle groups are commonly used, intramuscular injections may be given into the abdominals, biceps, calves, deltoids, gluteals, laterals, pectorals, quadriceps, trapezoids, and triceps.

Did you know?

In the United States, there is a birth every 8 seconds, according to the U.S. Census Bureau's Population Clock.

Did you know?

Although the Roman numeral for the number 4 has always been taught to have been "IV," according to historians, the ancient Romans probably used "IIII" most of the time. This is partially backed up by the fact that early grandfather clocks displayed IIII for the number 4 instead of IV. Early clockmakers apparently thought that the IIII balanced out the VIII (used for the number 8) on the clock face and that it just looked better.

For a complete list of videos, visit our video library