Author Question: The therapist is about to determine the respiratory system compliance of a patient receiving ... (Read 83 times)

oliviahorn72

  • Hero Member
  • *****
  • Posts: 579
The therapist is about to determine the respiratory system compliance of a patient receiving mechanical ventilation. The patient's tidal volume is 700 ml. Based on the pressure-time scalar presented below, which of the following pressures or pressure gradients does the therapist need to use in this calculation?
 
  A. The pressure reading at f needs to be used.
  B. The pressure gradient from g to f must be selected.
  C. The g-h pressure gradient must be chosen.
  D. The pressure measured at point g needs to be taken.

Question 2

While monitoring a mechanically ventilated patient, the therapist performs an expiratory hold maneuver to ascertain the presence of auto-PEEP. During that maneuver, the therapist notes that the pressure indicates 0 cm H2O. How should the therapist interpret this finding? I. The patient has no auto-PEEP. II. Airway pressure at this time equals alveolar pressure. III. The patient's static compliance has increased. IV. The patient's lungs have not completely emptied.
 
  A. I only
  B. IV only
  C. I, II only
  D. II, III only



choc0chan

  • Sr. Member
  • ****
  • Posts: 362
Answer to Question 1

ANS: C
A. Incorrect response: Label f on the pressure-time tracing represents the peak inspiratory pressure. This pressure is used in the denominator (minus total PEEP) of the formula used to calculate the dynamic compliance.
B. Incorrect response: The pressure gradient f minus g is the pressure generated during VT delivery required to overcome airway resistance. Specifically, these labels signify the difference between the peak inspiratory pressure minus the plateau pressure.
C. Correct response: The pressure gradient required for the calculation of respiratory system compliance is plateau pressure minus total PEEP. On the pressure-time scalar here, the gradient is g minus h.
D. Incorrect response: The pressure measure at g is the plateau pressure. The plateau pressure is used in the denominator (minus total PEEP) of the formula used to calculate the static compliance.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: During an expiratory hold maneuver, the flow during the expiratory phase is held at zero. When no gas flows at this time, alveolar pressure equilibrates with airway pressure. Therefore, if the reading during an expiratory hold is 0 cm H2O, the conclusion is that auto-PEEP is absent. Similarly, if this pressure measurement indicates some level of positive pressure, then auto-PEEP is present, assuming that the measured pressure exceeds any applied PEEP. In other words, if applied PEEP is 5 cm H2O, and the measured pressure during an expiratory hold is 10 cm H2O, 5 cm H2O of auto-PEEP is present.
D. Incorrect response: See explanation C.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
 

Did you know?

In 2012, nearly 24 milliion Americans, aged 12 and older, had abused an illicit drug, according to the National Institute on Drug Abuse (NIDA).

Did you know?

All patients with hyperparathyroidism will develop osteoporosis. The parathyroid glands maintain blood calcium within the normal range. All patients with this disease will continue to lose calcium from their bones every day, and there is no way to prevent the development of osteoporosis as a result.

Did you know?

Looking at the sun may not only cause headache and distort your vision temporarily, but it can also cause permanent eye damage. Any exposure to sunlight adds to the cumulative effects of ultraviolet (UV) radiation on your eyes. UV exposure has been linked to eye disorders such as macular degeneration, solar retinitis, and corneal dystrophies.

Did you know?

Most strokes are caused when blood clots move to a blood vessel in the brain and block blood flow to that area. Thrombolytic therapy can be used to dissolve the clot quickly. If given within 3 hours of the first stroke symptoms, this therapy can help limit stroke damage and disability.

Did you know?

Drug abusers experience the following scenario: The pleasure given by their drug (or drugs) of choice is so strong that it is difficult to eradicate even after years of staying away from the substances involved. Certain triggers may cause a drug abuser to relapse. Research shows that long-term drug abuse results in significant changes in brain function that persist long after an individual stops using drugs. It is most important to realize that the same is true of not just illegal substances but alcohol and tobacco as well.

For a complete list of videos, visit our video library