Author Question: Which maneuver will maintain air in the lungs at the end of inspiration, before the exhalation valve ... (Read 40 times)

Chelseaamend

  • Hero Member
  • *****
  • Posts: 545
Which maneuver will maintain air in the lungs at the end of inspiration, before the exhalation valve opens?
 
  a. Pressure limit
  b. Inspiratory hold
  c. Expiratory hold
  d. Expiratory resistance

Question 2

The respiratory therapist is called to a patient's room because the alarms are ringing. When the respiratory therapist arrives at the bedside, the high pressure limit, low exhaled tidal volume, and low exhaled minute volume alarms are active.
 
  The cause of these alarms is which of the following?
  a. Disconnection from the ventilator.
  b. Critical leak in the ventilator circuit.
  c. Lung compliance has improved.
  d. Airway resistance has increased.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

Ahnyah

  • Sr. Member
  • ****
  • Posts: 336
Answer to Question 1

ANS: B
The inspiratory hold, inspiratory pause, or end-inspiratory pause is the maneuver that will main-tain air in the lungs and extend inspiration. Pressure limit allows pressure to rise but not exceed a pressure limit setting. Expiratory hold is the maneuver that will obtain the unintended positive end expiratory pressure (auto-PEEP) measurement. The ventilator pauses before delivering the next machine breath. Expiratory resistance is a resistance added to exhalation to mimic pursed-lip breathing.

Answer to Question 2

ANS: D
The low exhaled tidal volume and minute volume alarms are active when the high pressure limit alarm is active. This occurs because reaching the set high pressure limit setting will end inspiration immediately by pressure cycling and thereby will decrease the volume delivered to the patient. When the high pressure limit alarm is active for several breaths, the low exhaled tidal volume and then the minute ventilation alarms will become active. The high pressure alarm will sound when airway resistance is elevated (e.g., asthma). A disconnect from the ventilator or a critical leak would cause the low peak inspiratory pressure alarms to ring. Improved lung compliance will lower the peak inspiratory pressure and may trigger a low pressure alarm.




Chelseaamend

  • Member
  • Posts: 545
Reply 2 on: Jul 16, 2018
Gracias!


at

  • Member
  • Posts: 359
Reply 3 on: Yesterday
Wow, this really help

 

Did you know?

Approximately one in three babies in the United States is now delivered by cesarean section. The number of cesarean sections in the United States has risen 46% since 1996.

Did you know?

The first monoclonal antibodies were made exclusively from mouse cells. Some are now fully human, which means they are likely to be safer and may be more effective than older monoclonal antibodies.

Did you know?

Fatal fungal infections may be able to resist newer antifungal drugs. Globally, fungal infections are often fatal due to the lack of access to multiple antifungals, which may be required to be utilized in combination. Single antifungals may not be enough to stop a fungal infection from causing the death of a patient.

Did you know?

According to animal studies, the typical American diet is damaging to the liver and may result in allergies, low energy, digestive problems, and a lack of ability to detoxify harmful substances.

Did you know?

Thyroid conditions may make getting pregnant impossible.

For a complete list of videos, visit our video library