Answer to Question 1
ANS: D
In patient with relatively healthy lungs where the termination criteria is set too high or inspiratory time too short, double triggering can occur every breath. These are primarily postoperative pa-tients or overdose patients.
Answer to Question 2
ANS: A
Normally the trigger delay should be minimal, less than 100 msec. When it exceeds 150 msec, the cause should be determined. Adjusting the sensitivity, setting the tidal volume appropriately and/or applying PEEP should correct delayed triggering unless there is a true malfunction of the ventilator.