Author Question: When the Drger Babylog 8000+ is used in the pressure support mode, the patient's peak inspir-atory ... (Read 84 times)

SO00

  • Hero Member
  • *****
  • Posts: 568
When the Drger Babylog 8000+ is used in the pressure support mode, the patient's peak inspir-atory flow is 30 L/min. At what flow rate will the ventilator cycle to expiration?
 
  a. 3.0 L/min c. 6.0 L/min
  b. 4.5 L/min d. 7.0 L/min

Question 2

On a Drger Babylog 8000, which of the following variables can trigger mandatory breaths in the synchronized intermittent mandatory ventilation mode?
 
  a. Volume and flow c. Time and volume
  b. Pressure and flow d. Pressure and time



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

sultana.d

  • Sr. Member
  • ****
  • Posts: 320
Answer to Question 1

ANS: B
Pressure support breaths are flow-cycled on the Drger Babylog 8000+ at 15 of peak flow.

Answer to Question 2

ANS: C
The mandatory breaths in the synchronized intermittent mandatory ventilation mode on the Dr-ger Babylog 8000 are either time-triggered or patient-triggered. Patient-triggering is accom-plished by volume, when a patient's spontaneous inspiratory volume is equal to or greater than the set trigger volume.





 

Did you know?

By definition, when a medication is administered intravenously, its bioavailability is 100%.

Did you know?

Anesthesia awareness is a potentially disturbing adverse effect wherein patients who have been paralyzed with muscle relaxants may awaken. They may be aware of their surroundings but unable to communicate or move. Neurologic monitoring equipment that helps to more closely check the patient's anesthesia stages is now available to avoid the occurrence of anesthesia awareness.

Did you know?

About 3.2 billion people, nearly half the world population, are at risk for malaria. In 2015, there are about 214 million malaria cases and an estimated 438,000 malaria deaths.

Did you know?

In 1844, Charles Goodyear obtained the first patent for a rubber condom.

Did you know?

Green tea is able to stop the scent of garlic or onion from causing bad breath.

For a complete list of videos, visit our video library