Author Question: A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, ... (Read 50 times)

abc

  • Hero Member
  • *****
  • Posts: 543
A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, the high-pressure alarm, which is set at 40 cm H2O, sounds periodically. What should the therapist do at this time?
 
  A. Increase the pressure limit.
  B. Change the Fio2.
  C. Increase the inspiratory flow.
  D. Continue ventilating the patient with the current settings.

Question 2

The therapist is using an O2-powered breathing device that is being manually triggered. Each time the therapist depresses the actuator, he continues to depress the actuator. What effect will maintaining pressure on the actuator have on the inspiratory time?
 
  A. It will have no influence on the inspiratory time.
  B. It will prolong the inspiratory time because inspiration will be maintained as long as the actuator is depressed.
  C. Inspiration will automatically be terminated after 3 seconds despite depressing the actuator continuously.
  D. Maintaining pressure on the actuator will cause multiple inspirations to be triggered.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

macagn

  • Sr. Member
  • ****
  • Posts: 345
Answer to Question 1

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: Transport ventilators contain a high-pressure relief valve, which vents gas to the atmosphere at a preselected peak inspiratory pressure. The pressure limit typically is 60 to 80 cm H2O for adults and 30 to 40 cm H2O for children. Activation of the high-pressure limit is usually signaled by a visual or audible alarm to alert the operator. Since the patient in this problem is periodically reaching the high pressure limit set at only 40 cm H2O, ventilation should continue as selected.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: O2-powered breathing devices (OPD) are frequently called demand valves. A typical OPD consists of a demand valve that can be manually triggered or patient triggered. Newer OPDs limit the available airway pressure via a pressure-limiting valve, and limit the length of the inspiratory time. In terms of the inspiratory time, inspiration is terminated after 3 seconds elapse, even if the operator continues to depress the actuator. Therefore maintaining pressure on the actuator for a prolonged period will have no influence on the inspiratory time.
D. Incorrect response: See explanation C




abc

  • Member
  • Posts: 543
Reply 2 on: Jul 16, 2018
Wow, this really help


kalskdjl1212

  • Member
  • Posts: 353
Reply 3 on: Yesterday
Excellent

 

Did you know?

People about to have surgery must tell their health care providers about all supplements they take.

Did you know?

Tobacco depletes the body of vitamins A, C, and E, which can result in any of the following: dry hair, dry skin, dry eyes, poor growth, night blindness, abscesses, insomnia, fatigue, reproductive system problems, sinusitis, pneumonia, frequent respiratory problems, skin disorders, weight loss, rickets, osteomalacia, nervousness, muscle spasms, leg cramps, extremity numbness, bone malformations, decayed teeth, difficulty in walking, irritability, restlessness, profuse sweating, increased uric acid (gout), joint damage, damaged red blood cells, destruction of nerves, infertility, miscarriage, and many types of cancer.

Did you know?

Colchicine is a highly poisonous alkaloid originally extracted from a type of saffron plant that is used mainly to treat gout.

Did you know?

The cure for trichomoniasis is easy as long as the patient does not drink alcoholic beverages for 24 hours. Just a single dose of medication is needed to rid the body of the disease. However, without proper precautions, an individual may contract the disease repeatedly. In fact, most people develop trichomoniasis again within three months of their last treatment.

Did you know?

Before a vaccine is licensed in the USA, the Food and Drug Administration (FDA) reviews it for safety and effectiveness. The CDC then reviews all studies again, as well as the American Academy of Pediatrics and the American Academy of Family Physicians. Every lot of vaccine is tested before administration to the public, and the FDA regularly inspects vaccine manufacturers' facilities.

For a complete list of videos, visit our video library