Author Question: A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, ... (Read 53 times)

abc

  • Hero Member
  • *****
  • Posts: 543
A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, the high-pressure alarm, which is set at 40 cm H2O, sounds periodically. What should the therapist do at this time?
 
  A. Increase the pressure limit.
  B. Change the Fio2.
  C. Increase the inspiratory flow.
  D. Continue ventilating the patient with the current settings.

Question 2

The therapist is using an O2-powered breathing device that is being manually triggered. Each time the therapist depresses the actuator, he continues to depress the actuator. What effect will maintaining pressure on the actuator have on the inspiratory time?
 
  A. It will have no influence on the inspiratory time.
  B. It will prolong the inspiratory time because inspiration will be maintained as long as the actuator is depressed.
  C. Inspiration will automatically be terminated after 3 seconds despite depressing the actuator continuously.
  D. Maintaining pressure on the actuator will cause multiple inspirations to be triggered.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

macagn

  • Sr. Member
  • ****
  • Posts: 345
Answer to Question 1

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: Transport ventilators contain a high-pressure relief valve, which vents gas to the atmosphere at a preselected peak inspiratory pressure. The pressure limit typically is 60 to 80 cm H2O for adults and 30 to 40 cm H2O for children. Activation of the high-pressure limit is usually signaled by a visual or audible alarm to alert the operator. Since the patient in this problem is periodically reaching the high pressure limit set at only 40 cm H2O, ventilation should continue as selected.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: O2-powered breathing devices (OPD) are frequently called demand valves. A typical OPD consists of a demand valve that can be manually triggered or patient triggered. Newer OPDs limit the available airway pressure via a pressure-limiting valve, and limit the length of the inspiratory time. In terms of the inspiratory time, inspiration is terminated after 3 seconds elapse, even if the operator continues to depress the actuator. Therefore maintaining pressure on the actuator for a prolonged period will have no influence on the inspiratory time.
D. Incorrect response: See explanation C




abc

  • Member
  • Posts: 543
Reply 2 on: Jul 16, 2018
Wow, this really help


momolu

  • Member
  • Posts: 320
Reply 3 on: Yesterday
Excellent

 

Did you know?

Essential fatty acids have been shown to be effective against ulcers, asthma, dental cavities, and skin disorders such as acne.

Did you know?

Coca-Cola originally used coca leaves and caffeine from the African kola nut. It was advertised as a therapeutic agent and "pickerupper." Eventually, its formulation was changed, and the coca leaves were removed because of the effects of regulation on cocaine-related products.

Did you know?

The first-known contraceptive was crocodile dung, used in Egypt in 2000 BC. Condoms were also reportedly used, made of animal bladders or intestines.

Did you know?

Approximately 25% of all reported medication errors result from some kind of name confusion.

Did you know?

The oldest recorded age was 122. Madame Jeanne Calment was born in France in 1875 and died in 1997. She was a vegetarian and loved olive oil, port wine, and chocolate.

For a complete list of videos, visit our video library