Author Question: A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, ... (Read 99 times)

abc

  • Hero Member
  • *****
  • Posts: 543
A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, the high-pressure alarm, which is set at 40 cm H2O, sounds periodically. What should the therapist do at this time?
 
  A. Increase the pressure limit.
  B. Change the Fio2.
  C. Increase the inspiratory flow.
  D. Continue ventilating the patient with the current settings.

Question 2

The therapist is using an O2-powered breathing device that is being manually triggered. Each time the therapist depresses the actuator, he continues to depress the actuator. What effect will maintaining pressure on the actuator have on the inspiratory time?
 
  A. It will have no influence on the inspiratory time.
  B. It will prolong the inspiratory time because inspiration will be maintained as long as the actuator is depressed.
  C. Inspiration will automatically be terminated after 3 seconds despite depressing the actuator continuously.
  D. Maintaining pressure on the actuator will cause multiple inspirations to be triggered.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

macagn

  • Sr. Member
  • ****
  • Posts: 345
Answer to Question 1

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: Transport ventilators contain a high-pressure relief valve, which vents gas to the atmosphere at a preselected peak inspiratory pressure. The pressure limit typically is 60 to 80 cm H2O for adults and 30 to 40 cm H2O for children. Activation of the high-pressure limit is usually signaled by a visual or audible alarm to alert the operator. Since the patient in this problem is periodically reaching the high pressure limit set at only 40 cm H2O, ventilation should continue as selected.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: O2-powered breathing devices (OPD) are frequently called demand valves. A typical OPD consists of a demand valve that can be manually triggered or patient triggered. Newer OPDs limit the available airway pressure via a pressure-limiting valve, and limit the length of the inspiratory time. In terms of the inspiratory time, inspiration is terminated after 3 seconds elapse, even if the operator continues to depress the actuator. Therefore maintaining pressure on the actuator for a prolonged period will have no influence on the inspiratory time.
D. Incorrect response: See explanation C




abc

  • Member
  • Posts: 543
Reply 2 on: Jul 16, 2018
:D TYSM


DylanD1323

  • Member
  • Posts: 314
Reply 3 on: Yesterday
Gracias!

 

Did you know?

More than one-third of adult Americans are obese. Diseases that kill the largest number of people annually, such as heart disease, cancer, diabetes, stroke, and hypertension, can be attributed to diet.

Did you know?

Studies show that systolic blood pressure can be significantly lowered by taking statins. In fact, the higher the patient's baseline blood pressure, the greater the effect of statins on his or her blood pressure.

Did you know?

After a vasectomy, it takes about 12 ejaculations to clear out sperm that were already beyond the blocked area.

Did you know?

Common abbreviations that cause medication errors include U (unit), mg (milligram), QD (every day), SC (subcutaneous), TIW (three times per week), D/C (discharge or discontinue), HS (at bedtime or "hours of sleep"), cc (cubic centimeters), and AU (each ear).

Did you know?

Before a vaccine is licensed in the USA, the Food and Drug Administration (FDA) reviews it for safety and effectiveness. The CDC then reviews all studies again, as well as the American Academy of Pediatrics and the American Academy of Family Physicians. Every lot of vaccine is tested before administration to the public, and the FDA regularly inspects vaccine manufacturers' facilities.

For a complete list of videos, visit our video library