Author Question: A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, ... (Read 98 times)

abc

  • Hero Member
  • *****
  • Posts: 543
A patient is being ventilated by a transport ventilator en route to a hospital. During ventilation, the high-pressure alarm, which is set at 40 cm H2O, sounds periodically. What should the therapist do at this time?
 
  A. Increase the pressure limit.
  B. Change the Fio2.
  C. Increase the inspiratory flow.
  D. Continue ventilating the patient with the current settings.

Question 2

The therapist is using an O2-powered breathing device that is being manually triggered. Each time the therapist depresses the actuator, he continues to depress the actuator. What effect will maintaining pressure on the actuator have on the inspiratory time?
 
  A. It will have no influence on the inspiratory time.
  B. It will prolong the inspiratory time because inspiration will be maintained as long as the actuator is depressed.
  C. Inspiration will automatically be terminated after 3 seconds despite depressing the actuator continuously.
  D. Maintaining pressure on the actuator will cause multiple inspirations to be triggered.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

macagn

  • Sr. Member
  • ****
  • Posts: 345
Answer to Question 1

ANS: D
A. Incorrect response: See explanation D.
B. Incorrect response: See explanation D.
C. Incorrect response: See explanation D.
D. Correct response: Transport ventilators contain a high-pressure relief valve, which vents gas to the atmosphere at a preselected peak inspiratory pressure. The pressure limit typically is 60 to 80 cm H2O for adults and 30 to 40 cm H2O for children. Activation of the high-pressure limit is usually signaled by a visual or audible alarm to alert the operator. Since the patient in this problem is periodically reaching the high pressure limit set at only 40 cm H2O, ventilation should continue as selected.

Answer to Question 2

ANS: C
A. Incorrect response: See explanation C.
B. Incorrect response: See explanation C.
C. Correct response: O2-powered breathing devices (OPD) are frequently called demand valves. A typical OPD consists of a demand valve that can be manually triggered or patient triggered. Newer OPDs limit the available airway pressure via a pressure-limiting valve, and limit the length of the inspiratory time. In terms of the inspiratory time, inspiration is terminated after 3 seconds elapse, even if the operator continues to depress the actuator. Therefore maintaining pressure on the actuator for a prolonged period will have no influence on the inspiratory time.
D. Incorrect response: See explanation C




abc

  • Member
  • Posts: 543
Reply 2 on: Jul 16, 2018
:D TYSM


AISCAMPING

  • Member
  • Posts: 347
Reply 3 on: Yesterday
Thanks for the timely response, appreciate it

 

Did you know?

In most climates, 8 to 10 glasses of water per day is recommended for adults. The best indicator for adequate fluid intake is frequent, clear urination.

Did you know?

The modern decimal position system was the invention of the Hindus (around 800 AD), involving the placing of numerals to indicate their value (units, tens, hundreds, and so on).

Did you know?

After 5 years of being diagnosed with rheumatoid arthritis, one every three patients will no longer be able to work.

Did you know?

The human body produces and destroys 15 million blood cells every second.

Did you know?

All adults should have their cholesterol levels checked once every 5 years. During 2009–2010, 69.4% of Americans age 20 and older reported having their cholesterol checked within the last five years.

For a complete list of videos, visit our video library