Author Question: The therapist is to manually calculate expiratory airway resistance on a patient receiving ... (Read 42 times)

chads108

  • Hero Member
  • *****
  • Posts: 507
The therapist is to manually calculate expiratory airway resistance on a patient receiving mechanical ventilation. What should the therapist do to ensure valid measurements? I. Set the ventilator to deliver a descending ramp flow pattern II. Initiate an expiratory hold maneuver at the end of expiration. III. Ensure that the patient receives a mandatory breath. IV. Institute an inspiratory pause maneuver at the end of inspiration.
 
  A. I, II only
  B. III, IV only
  C. I, II, III only
  D. II, III, IV only

Question 2

A patient receiving mechanical ventilation is about to receive medication delivered via an MDI. Where should the therapist place the spacer, and when should the MDI be actuated?
 
  A. Place the spacer anywhere along the inspiratory limb, and actuate the MDI at the end of inspiration.
  B. Place the spacer in the inspiratory side of the ventilator circuit at the Y, and actuate the MDI at end-exhalation.
  C. Place the spacer near the elbow connector, and actuate the MDI anytime during the ventilatory cycle.
  D. Place the spacer along the inspiratory limb about 50 cm from the elbow connector, and actuate the MDI during inspiration.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question
Marked as best answer by a Subject Expert

dmurph1496

  • Sr. Member
  • ****
  • Posts: 343
Answer to Question 1

ANS: B
A. Incorrect response: See explanation B.
B. Correct response: The ventilator must be set to deliver a square wave inspiratory flow pattern. Any other flow pattern will provide invalid data. If the patient contributes any inspiratory or expiratory effort, the measurement is invalid. The delivered breath must be a mandatory breath.
An expiratory pause is used to measure auto-PEEP, not expiratory airway resistance. An inspiratory pause is initiated at the end of inspiration to obtain peak pressure and plateau pressure readings. Therefore, the pressure generated to overcome airway resistance (driving pressure) can be determined.
C. Incorrect response: See explanation B.
D. Incorrect response: See explanation B.

Answer to Question 2

ANS: B
A. Incorrect response: See explanation B.
B. Correct response: When an MDI is used to deliver medication to a mechanically ventilated patient, the spacer should be attached to the Y in the inspiratory side of the circuit. The MDI can be actuated at end-exhalation so that the drug plume is inspired during the next inspiration. The MDI can also be actuated in synchrony with inspiration.
C. Incorrect response: See explanation B.
D. Incorrect response: See explanation B.





 

Did you know?

Computer programs are available that crosscheck a new drug's possible trade name with all other trade names currently available. These programs detect dangerous similarities between names and alert the manufacturer of the drug.

Did you know?

Liver spots have nothing whatsoever to do with the liver. They are a type of freckles commonly seen in older adults who have been out in the sun without sufficient sunscreen.

Did you know?

Looking at the sun may not only cause headache and distort your vision temporarily, but it can also cause permanent eye damage. Any exposure to sunlight adds to the cumulative effects of ultraviolet (UV) radiation on your eyes. UV exposure has been linked to eye disorders such as macular degeneration, solar retinitis, and corneal dystrophies.

Did you know?

Thyroid conditions may make getting pregnant impossible.

Did you know?

Oliver Wendell Holmes is credited with introducing the words "anesthesia" and "anesthetic" into the English language in 1846.

For a complete list of videos, visit our video library