Author Question: When adaptive support ventilation (ASV) is being used, how does the Hamilton Galileo minimize air ... (Read 70 times)

joesmith1212

  • Hero Member
  • *****
  • Posts: 549
When adaptive support ventilation (ASV) is being used, how does the Hamilton Galileo minimize air trapping and auto-PEEP?
 
  A. It automatically decreases the exiting VT when necessary to prevent overdistention of the lungs.
  B. The ventilator adjusts the flow-cycle criterion on a breath-to-breath basis.
  C. It calculates the expiratory time constant and uses the time constant to maintain adequate expiratory time.
  D. The ventilator accomplishes this by either increasing or decreasing pressure support.

Question 2

18. While mechanically ventilating a patient with VC-SIMV with PSV of 10 cm H2O, the therapist seeks to increase the patient's spontaneous VT from 150 ml to 300 ml. How can the therapist accomplish this task?
 
  A. Switch to PC-SIMV with PSV.
  B. Increase the minute ventilation control.
  C. Increase the inspiratory rise time.
  D. Adjust the flow termination criterion from 40 to 25.



Ksanderson1296

  • Sr. Member
  • ****
  • Posts: 339
Answer to Question 1

ANS: C
A. Incorrect response: During ASV, the ventilator (1) uses the initial ventilator settings (i.e., high-pressure alarm, PEEP, and Fio2), (2) calculates system compliance, airway resistance, and auto-PEEP, and (3) uses the pressure limit of both mandatory and spontaneous breaths to determine the target breath and tidal volume.
B. Incorrect response: The therapist can adjust the rise time and the flow-cycling criterion.
C. Correct answer: The ventilator adjusts the I:E ratio and the inspiratory time of the mandatory breaths to prevent air trapping. These adjustments are accomplished by calculating the patient's expiratory time constants, and maintaining sufficient expiratory time.
D. Incorrect response: Pressure support ventilation is not part of ASV.

Answer to Question 2

ANS: D
A. Incorrect response: Switching to PC-SIMV with PSV would have no influence on the magnitude of the patient's spontaneous tidal volume. All that would be accomplished by making this switch is substituting volume-control ventilation for pressure-control ventilation. This change would affect only the mandatory breaths.
B. Incorrect response: MMV enables the patient to breath spontaneously, and ensures that the patient receives a minimum minute ventilation initially set by the therapist. However, MMV is a stand-alone mode and is not used with PC-SIMV with PSV.
C. Incorrect response: Increasing the inspiratory rise time has the effect of lengthening the time required for the pressure to reach its target value. Increasing the inspiratory rise time inappropriately can produce an increased work of breathing for the patient. Conversely, decreasing the inspiratory rise time too much can cause pressure overshoot, which contributes to premature cycling.
D. Correct response: Changing the flow-cycle criterion from 40 to 25 during PSV lengthens the inspiratory time and increases the tidal volume. Conversely, if the flow-cycle criterion is changed from 5 to 50, for example, the inspiratory time shortens and the tidal volume decreases. For ventilators lacking the flow-cycle criterion function, the therapist will increase the PSV level.



Related Topics

Need homework help now?

Ask unlimited questions for free

Ask a Question


 

Did you know?

When intravenous medications are involved in adverse drug events, their harmful effects may occur more rapidly, and be more severe than errors with oral medications. This is due to the direct administration into the bloodstream.

Did you know?

Colchicine is a highly poisonous alkaloid originally extracted from a type of saffron plant that is used mainly to treat gout.

Did you know?

In the United States, congenital cytomegalovirus causes one child to become disabled almost every hour. CMV is the leading preventable viral cause of development disability in newborns. These disabilities include hearing or vision loss, and cerebral palsy.

Did you know?

The average human gut is home to perhaps 500 to 1,000 different species of bacteria.

Did you know?

During the twentieth century, a variant of the metric system was used in Russia and France in which the base unit of mass was the tonne. Instead of kilograms, this system used millitonnes (mt).

For a complete list of videos, visit our video library